Introduction to X-Ray Emission Spectrometry

7.3 Detection limits


Detection limit (DL) is the lowest amount of an analyte to be examined in a test material that can be detected and regarded as different from the blank value (with a given probability), but not necessarily quantified. In fact, two risks must be taken into account: false positives (the risk α of considering the element is present in test material when its quantity is null) and; false negatives (the risk β of considering a substance is absent from a substance when its quantity is not null).

Detection limits are important for several reasons:

The evaluation of the detection limits requires a careful assessment, based on statistics approach and on the specifics of EDXRF analysis.

The detection limits are usually expressed in units of calculated weight fraction (or grams per square centimetre). The latter is achieved by evaluating the noise counts in the formula used to calculate the weight fraction (or areal deposit).

(for thin samples)

(for thick samples)

As the measured counts N depend on the time of measurement (N = I × tmeas) the detection limits are directly proportional to the square root of the noise, inversely proportional to the instrumental sensitivity S and inversely proportional to the square root of the measurement time tmeas.

The noise signal in EDXRF can include different non-correlated contributions. In the more general case, its combined standard deviation can be expressed using the rules for error propagation:

where:

Reference: R. Padilla Álvarez, D. Hernández Torres, A. Markowicz, D. Wregzynek, E. Chinea Cano, S. A. Bamford, Quality management and method validation in EDXRF analysis, X-Ray Spectrometry 36 (2007) 27.