Introduction to X-Ray Emission Spectrometry

7.2 Working interval


The measuring interval (sometimes called working range) is the set of values of quantities of the same kind that can be measured by a given instrument (or measuring system) with specified uncertainty under defined conditions.

In most of the analytical chemistry methods, the working interval is defined as that for which the linearity in the calibration is maintained. As mentioned in Section 6, linearity is only observed in a few specific EDXRF methods. In such cases the interval of concentrations for which the linearity is fulfilled must be assessed. The latter can be done by any of the two following approaches:

For other EDXRF methods, the interval of applicability of the model has to be assessed based on the assumptions made for the given calibration model. In the case of thin sample analysis the increase in concentration of an element can enlarge the χS(E0,Ei)ρX value. For low characteristic energies the value of χS(E0,Ei) must be verified for the given sample matrix and sample areal density. The presence of high concentration elements in the sample might require the use of additional corrections for enhancement effects in fundamental parameter based methods.

References:

ISO 11095:1996. Linear calibration using reference materials.